Greetings,

Anyone out there have experience with aldrin recovery problems in EPA 525? We seem to have trouble in spiked water samples where aldrin is at low level while other compounds on the list are at high levels that require additional dilution. It is the primary run, without dilution, where aldrin fails recovery with low bias.

This is frustrating since aldrin is fine in our QC, including a low level laboratory control sample.

Anyone out had to deal with aldrin?

TIA,

Evan L. Cooper

I have some evidence of low aldrin recoveries out of samples but not consistent enough to track a hard and fast problem. We run 525.2 SPE on mostly clean drinking waters. Typically, %R @ 1ug/l spike recovers at >90%. Some sample spikes can be as low as 55%.

BTW - We have never had a need to dilute a sample for high contaminants. How do you dilute an extract, as the IS is added prior to extraction?

The dilutions I mentioned were performed prior to the extraction.

What SPE are you using?

ravenwork

We use the C18 EMPORE extraction disks.
4 posts Page 1 of 1

Who is online

In total there are 228 users online :: 1 registered, 0 hidden and 227 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Ahrefs [Bot] and 227 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry