8260B Low Level Soil Analysis
Posted: Fri Sep 09, 2016 2:34 am
Hi all! I'm having some issues and I was hoping you could provide some guidance!
I'm running an Agilent 7890/5975 GC/MSD with a Centurion autosampler/Evolution concentrator. I'm looking for help with the 8260B low level soil analysis. When running water and methanol extracted soil dilutions I have no problems. However, for the LL soils I am having problems with low internal standard recoveries (chlorobenzene-d5 specifically) in samples only. My theory is that it has something to do with the coelution of the deuterated and non-deuterated compounds in the calibration standards. For those of you who are not familiar, internal standard area recoveries need to be 50-200% of the area of the same internal standard in the continuing calibration standard. My thought is that when the calibration is analyzed (the internal standard mix is spiked during sampling by the autosampler) the coelution causes elevated areas for this compound. The same thing occurs with the CCV (and any spike). Then when the samples and method blank are analyzed the absence of the chlorobenzene from the calibration mix causes the recovery to be too low to meet acceptance criteria. This then often affects surrogate recovery as well.
Again, this occurance is strange to me because my system does just fine when using the sparge tube to purge waters or methanolic dilutions. So, is it a problem with the autosampler purge (since it purges in the vial itself)? The line from the purge vessel to the trap is longer than it would be for waters analysis but the desorb would be the same. Or is it the GC oven temp program? Or the MSD?
Any help would be greatly appreciated!
Thanks!
Just a note: I was actually hired to help fix the mess that was left by the previous manager of the volatiles department at my lab. I do not know what was done in the past or if they had the same issues that I'm having. My background is in 8020/21B which is PID/FID so I am new-ish to MS but am pretty fluent with P&T/GC.
Carrie
I'm running an Agilent 7890/5975 GC/MSD with a Centurion autosampler/Evolution concentrator. I'm looking for help with the 8260B low level soil analysis. When running water and methanol extracted soil dilutions I have no problems. However, for the LL soils I am having problems with low internal standard recoveries (chlorobenzene-d5 specifically) in samples only. My theory is that it has something to do with the coelution of the deuterated and non-deuterated compounds in the calibration standards. For those of you who are not familiar, internal standard area recoveries need to be 50-200% of the area of the same internal standard in the continuing calibration standard. My thought is that when the calibration is analyzed (the internal standard mix is spiked during sampling by the autosampler) the coelution causes elevated areas for this compound. The same thing occurs with the CCV (and any spike). Then when the samples and method blank are analyzed the absence of the chlorobenzene from the calibration mix causes the recovery to be too low to meet acceptance criteria. This then often affects surrogate recovery as well.
Again, this occurance is strange to me because my system does just fine when using the sparge tube to purge waters or methanolic dilutions. So, is it a problem with the autosampler purge (since it purges in the vial itself)? The line from the purge vessel to the trap is longer than it would be for waters analysis but the desorb would be the same. Or is it the GC oven temp program? Or the MSD?
Any help would be greatly appreciated!
Thanks!
Just a note: I was actually hired to help fix the mess that was left by the previous manager of the volatiles department at my lab. I do not know what was done in the past or if they had the same issues that I'm having. My background is in 8020/21B which is PID/FID so I am new-ish to MS but am pretty fluent with P&T/GC.
Carrie