EZChrom noise calculation
Posted: Tue Feb 28, 2006 1:27 pm
by JAROCZAJKA
Hello All,
I have another question about EZChrom software. Does anybody know how the S/N ratio is calculated during postprocessing. As I know S/N is calculated as height of the peak of interest divided by noise calculated in the reange between the peaks. But it seem to be crazy ? Don't you?
JaroCzajka
Posted: Mon Mar 06, 2006 9:31 pm
by unmgvar
you should go to the help in the software and look at the equations displayed there for the noise and S/N calculations.
they actually do comply.
one draw back is that in ezchrom you cannot calculate S/N if you have a peak elution during a gradient stage because the drift of the baseline in many case due to the changing solvent constitution will add a great margin of error.
Posted: Tue Mar 07, 2006 9:26 am
by JAROCZAJKA
It seems to me the best way is to do some more "scripting"
In my opinion I had to store blank chromatogram (base line) and use it as a referance for calculation noise according to the pharmacopea.
Re: S/N Calcs
Posted: Thu Mar 09, 2006 8:04 am
by jwmol
There are several noise calculation procedures. In Europe one that is applied often requires the noise form a blank run to be applied to the peak heights of the sample.
This specific one can only be calculated in EZChrom with a scripting calculation (custom parameter) where the parameter will store the noise value of the blank run in a temporary file and uses that value for all subsequent sample runs.