Dear friends,
I am looking to find a standard method for analaysing MTBE in water. I want just to find the methods which work with GC/FID .
Would you please guide me if there is a standard analitical method for identifying and measuring MTBE in water (ASTM is preferably needed).
Also please guide me if any other link related is.

Thank you in advance.
Yours faithfully,
Semiramis

Posted: Sat Feb 25, 2006 6:58 pm
by Noser222
I searched on Yahoo
This should help you:

www.epa.gov/ORD/NRMRL/Pubs/600JA00193/600ja00193.pdf

Analayzing MTBE in Water

Posted: Sun Feb 26, 2006 5:34 am
by chemengineer
Dear Noser,
Many thanks for your reply, but I want to find a method for analayzing MTBE in water with GC. I don't want to find a treatment method for MTBE removing, just want to measure MTBE concentrations in Water.

If you have any suggestion I thank you.
Sincerely ,
Semiramis

Posted: Sun Feb 26, 2006 12:28 pm
by rc_12321
Prabably the following link helps you.

http://ijims.ansci.de/pdf/5/3/18_Stach_ISIMS-02.pdf

Analayzing MTBE in Water / GC

Posted: Mon Feb 27, 2006 5:08 am
by chemengineer
Dear Sir/Madam,
Many thanks for your kind reply to my request.
Actually I need to do standard method for measuring MTBE concentrations in water and I have to apply just a GC instrument with FID detector so I dont have any IMS instrument.
If you know about analayzing by GC, could you please tell me about.

Thank you for your kind attentions.
Yours faithfully,
Semiramis Badri

MtBE in water

Posted: Mon Feb 27, 2006 1:06 pm
by chromatographer1
There are many ways of performing this test. The problem with any 'standard' method are the possible trace impurities in the water which may give you a false positive.

Headspace analysis will give you a 1ng per 100mg sensitivity, but how low do you want to go. I suspect a lower limit of detection is desired.

A good method can be developed by using two dimensional GC technology. Large injection technique could be implemented to increase sensitivity. Simply use a non-polar column as your backflush column and a polar column as your analytical column.

Purge and trap is probably the best documented technique. Try reviewing the California EPA web site for the methods they use.

best wishes,

Rod

MTBE in Water

Posted: Tue Feb 28, 2006 7:01 am
by chemengineer
Dear Rod,
Thank you for your help.

Posted: Wed Mar 01, 2006 8:57 pm
by Steve Reimer
http://www.epa.gov/epaoswer/hazwaste/te ... /8015b.pdf

Around here the standard is SW-846 method 8015b with method 5035 (purge and trap). It can be done along with the BETX components during a gasoline analysis.
Disclaimer: this may not be sensitive enough for regulatory purposes in all areas.