Agilent 7000 GC/MS
Posted: Mon Sep 05, 2016 3:08 pm
by Christian_bios
I have a problem with autotuning the MS.
Prior to the current error I had the problem, that the autotune resulted in an error message stating that it could not "maintain voltage".
After this I cleaned all the parts of the source. Now the previous problem is gone, but still the autotune fails. It seems I am not getting any signal... just seeing the background noise.
Can someone suggest possible solutions for the problem? Maybe some typical errors for the agilent 7000 series?
Re: Agilent 7000 GC/MS
Posted: Mon Sep 05, 2016 3:14 pm
by Rndirk
Do you see the calibration gas peak if you manually turn on the MS (and the calibration gas valve)?
The problem could be caused by:
- Filament broken
- Calibration gas empty or somehow doesn't release into the MS
- A huge leak in the MS, this can completely mask your tuning component
Posted: Tue Sep 06, 2016 6:22 am
by Christian_bios
Thanks for your fast reply and your suggestion. I do not see a calibration gas peak. I will check for the calibration gas and filament immediately.
I would not expect a huge leak in the MS, since the vacuum seems to be fine and the pump is operation at around 20W.
I will let you know soon how things work out...
Posted: Tue Sep 06, 2016 6:34 am
by Rndirk
Before the next autotune, try to get a normal signal in manual tune mode first, then do an air/water check.
If your EI source has 2 filaments, you can easily switch between them in the manual tune mode to check if it's the filament.
Also check out
http://www.agilent.com/cs/library/userm ... intain.pdf
page 46
Posted: Tue Sep 06, 2016 11:07 am
by Christian_bios
I exchanged the filament, and the problem is now back to "can not maintain voltage".
I am not sure what is meant by "get a normal signal in manual tune mode"
I have been working through the troubleshoot guide, but could not find any good ideas for a possible solution.
here a picture from error message
http://imgur.com/a/NhgtD
Posted: Wed Sep 07, 2016 2:38 pm
by Christian_bios
The problem was found: The Lens-Insulator was damaged. This caused a shortcut between the ion-focus lens and the source body via the setscrew.
Easy fix: Loosen the setscrew to stop the shortcut.
Long-term: replace lens-insulator