ATD Desorption Efficiency
Posted: Thu Sep 01, 2016 11:06 pm
Hi All,
I'm running into another issue with my PE Turbomatrix 650. I can't seem to get any desorption of the heavier compunds. I'm running this program from the ATD onto an HP 5890/5972.
Room temp dry purge for 6.0 minutes at 50 mL/min
desob1 (onto cold trap) 330 C 6 minutes at 40 mL/min
desob2 (onto column) trap heat from -20->325 99C/sec
1.2 mL/minute onto column flow 6 min desorb
i'm using the PE Air toxics trap and tubes. we're running EPA method TO-17, at least trying to.
Now here is the weird part. the 650 has the Internal standard injection option, which we use, i can get recoveries of IS up to 18 minutes into the 27 minute run but my analytes that were spiked onto the tube prior to loading stop around 12 minutes. We also have a purge and trap GC/MS set up so i ran the spike standard there and everything recovered perfectly. the concentration of spike doesn't seem to matter either, i have loaded the tubes at 400 ng/l and get identical results, as far which compounds recover, to the 50 ng/L spike. I've tried liquid spiking the tubes and gas phase spiking. at one point i jumped the stage 1 desorb, onto trap, flow rate to 70 mL/min all to no avail. I'm at Wits End, any input is appreciated.
I'm running into another issue with my PE Turbomatrix 650. I can't seem to get any desorption of the heavier compunds. I'm running this program from the ATD onto an HP 5890/5972.
Room temp dry purge for 6.0 minutes at 50 mL/min
desob1 (onto cold trap) 330 C 6 minutes at 40 mL/min
desob2 (onto column) trap heat from -20->325 99C/sec
1.2 mL/minute onto column flow 6 min desorb
i'm using the PE Air toxics trap and tubes. we're running EPA method TO-17, at least trying to.
Now here is the weird part. the 650 has the Internal standard injection option, which we use, i can get recoveries of IS up to 18 minutes into the 27 minute run but my analytes that were spiked onto the tube prior to loading stop around 12 minutes. We also have a purge and trap GC/MS set up so i ran the spike standard there and everything recovered perfectly. the concentration of spike doesn't seem to matter either, i have loaded the tubes at 400 ng/l and get identical results, as far which compounds recover, to the 50 ng/L spike. I've tried liquid spiking the tubes and gas phase spiking. at one point i jumped the stage 1 desorb, onto trap, flow rate to 70 mL/min all to no avail. I'm at Wits End, any input is appreciated.