Air bubbles after changing from binarty to quarternary pump?
Posted: Thu Sep 01, 2016 12:18 pm
Hello fellow Chromatographers,
I am confronted with problems regarding Air bubbles. Currently, I am using the Agilent 1100 system.
Degasing is done with the "Merck Solvent Degasser L-7612" used for LaChrom system, since we don't have the inline degaser from Agilent.
I use a NH2-column for HILIC mode with the aim of analysing glycoprofiles of glycoproteins.
The solvents I use are: 100% Acetonitrile and Ammonium formiate 250mM (pH 4.4) and Isopropanol for purging.
So, the story: The old binary pump leaked and was substituted for a quarternary pump. Using the binary pump, the pressure was constant. Now with the quarternary pump, the pressure fluctuates strongly. The air bubbles appear after the the 4-channel mixing chamber.
History:
binary pump + degasing with supersonic = ok
quarternary pump + degasing with supersonic = very bad
quarternary pump + solvent degaser = better, but still random air bubbles
A few observations:
Purging with isopropanol to waste or pumping it through the column results in constant pressure.
Purging with Acetonitrile to waste or through column = constant pressure.
Acetonitrile+Ammonium formate = gets worse after a while.
(Gradient: ACN from 75% down to 50% in 25minutes). Especially, at the time when ammonium formiate has a higher percentage, the pressure fluctuates very strong.
I assume that is the reason why my dextran standard had different retention times with 3 runs at the same day.
Does anyone know what kind of problem I am dealing with this ?
Broken chamber or problems with solvent B ? I can also upload the chromatograms if anyone is interested.
Thanks, best Regards
Moe
I am confronted with problems regarding Air bubbles. Currently, I am using the Agilent 1100 system.
Degasing is done with the "Merck Solvent Degasser L-7612" used for LaChrom system, since we don't have the inline degaser from Agilent.
I use a NH2-column for HILIC mode with the aim of analysing glycoprofiles of glycoproteins.
The solvents I use are: 100% Acetonitrile and Ammonium formiate 250mM (pH 4.4) and Isopropanol for purging.
So, the story: The old binary pump leaked and was substituted for a quarternary pump. Using the binary pump, the pressure was constant. Now with the quarternary pump, the pressure fluctuates strongly. The air bubbles appear after the the 4-channel mixing chamber.
History:
binary pump + degasing with supersonic = ok
quarternary pump + degasing with supersonic = very bad
quarternary pump + solvent degaser = better, but still random air bubbles
A few observations:
Purging with isopropanol to waste or pumping it through the column results in constant pressure.
Purging with Acetonitrile to waste or through column = constant pressure.
Acetonitrile+Ammonium formate = gets worse after a while.
(Gradient: ACN from 75% down to 50% in 25minutes). Especially, at the time when ammonium formiate has a higher percentage, the pressure fluctuates very strong.
I assume that is the reason why my dextran standard had different retention times with 3 runs at the same day.
Does anyone know what kind of problem I am dealing with this ?
Broken chamber or problems with solvent B ? I can also upload the chromatograms if anyone is interested.
Thanks, best Regards
Moe