I am trying to determine exactly what is happening from your post. If I misunderstand please forgive.
One question I have is WHEN this occurs. First injection of many ?
or after several injections have occurred?
The rise in baseline is probably the residual DMSO left in the inlet (placement of the needle in the injection port) or on the column (the 'tail' of the previous diluent peak injection).
One solution is to increase the time you 'cook off' the column at the end of the run. I would suggest you do not use 300°C for 2 minutes but 280°C for 10 min. and see if that helps to better elute the DMSO 'tail' from the column.
I would elute DMSO diluent from my HS columns at a much lower temperature but for a longer time when I did my HS work 10 years ago. I also directly hooked up the transfer line to my column instead of using a split injection.
Perhaps a reading of the article might be helpful.
Rapid Analysis of USP Organic Volatile Impurities and Thirteen Other Common Residual Solvents by Static Headspace Analysis
Journal of Analytical Chemistry 1997 June
best wishes,
Rodney George
Senior Research and Development Scientist
Gas Separations Research
Supelco