Hi everyone,

Im setting up our SIL-10A (autosampler) with the syringe and wondering if anyone knows where the right (and bottom) tubing go to?

That is, do I leave the right/bottom hole open, or ? Are these going to waste/drain?

The syringe and SIL are connected and communicating properly so no issue there.

Image
The right hand side port is connected to the needle inside the autosampler.

Bottom port , no connection. It is leakage port in case of rotor problem or overpressure.
Correction : the right hand side port is connected to washing solvent.
Other port is already connected to autosampler needle.
Thanks uzman for the reply and correction!

Can I use the same mobile phase (stored in a different bottle) as the 'washing solvent'?
Hello lespaul ,

Recommended washing solvent is iso-propanol but you may use other solvents which is best washing properties for the injected compounds.
5 posts Page 1 of 1

Who is online

In total there are 21 users online :: 3 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Baidu [Spider], Semrush [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry