Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hi, when trying to use isotope as inter. std, a interesting question come into my mind. Under MRM mode, if the parent ions of analyte and internal Std have different masses(e.g. isotope labled for the internal std), the daughter ions which was selected to monitor have same masses however(e.g. the selected fragment is not isotope labeled for the internal std). Could the instrument distinguish them and give correct analysis? Thanks.
emily lee

Normally you should be able to so as the collision cell is emptied before selecting the next ion for fragmentation. That said, there is always the problem of cross-talk that happens exactly at the cases you mention. There are solutions for that though, you can either use a higher pause time or maybe the best one is to use a phantom MRM transition between these two i.e. something with a parent and daughter ion totally unrelated with your analytes...

New generation mass spectrometers have smaller problems of cross-talk than older generations...
2 posts Page 1 of 1

Who is online

In total there are 192 users online :: 1 registered, 0 hidden and 191 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Amazon [Bot] and 191 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry