high noise level

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi
we use GC/FID shimadzu 14 A.
recently, we moved the instrument, now observing banormally high noise level.
I clean the injector, FID cell, collector,... but this problem dosent solve!
there are noise while the FID dont ignite.
can any one help me!!!!
thanks

solve my problem.
the earth dosnt connect.
thanks

You mean yiou have noise even with the detector's flame off?

If so, your problem is in the electrical installation. Ask an electricitian to check for electrical / electronical noise, if the earth is really 0 V and you do not have fluctuations in power supply.