- Posts: 38
- Joined: Fri Oct 28, 2005 1:44 pm
I want to inject hydrogen/air mixtures, either by syringe or directly from a small sampling cylinder at slightly elevated pressures (eg 30 psi), without introducing air into the inlet, since I need to analyze for air in the sample, or at least be sure that hydrogen isn't escaping as air is entering (air can then be determined by difference since the total pressure of the sample will be known). I've been told by chromatographer1 and others that the only way to avoid air contamination is to use a sampling valve.
Will this valve be ok for use with GC? I downloaded the operating instructions for the valve from the Rheodyne website, and it looks simple enough if I purchase a 5 or 10 uL sample loop (it has a 100 uL sample loop on it now) and inject 1-2 mL of sample with a syringe to adequately purge the sample through the loop. Valco actually recommended an HPLC valve to me over the phone, but this could have been an error. What's not clear is if this will actually correct the problem of air entering during the injection, not knowing if air is entering at the moment of injection or if it is continuously seeping in while the needle is in place. Any suggestions are appreciated.
Also, to get a more adequate purge, perhaps I could connect the waste line to a vacuum pump to be able to vacuum purge the sample loop?