- Posts: 25
- Joined: Tue Jun 07, 2016 8:16 am
I was using a Zebron 35HT (30x0,25<0,25) column for about 3 months before it started tailing on my GCMS System. I was using a deactivated guard column (2,5m x 0,25) which I had to shorten / exchange periodically to get acceptable tailing. I changed the liner about once a week due to increased (tolerable) tailing. however, after 3 months tailing wouldnt get better with that 35HT column even though I changed septum, liner, guard column and syringe. Even injecting directly on the 35HT didnt resolve the tailing problem. (Splitless directly on the analytical column doesnt work accordingly does it?)
So I tried using a 30x0,25x0,25 HP5ms which is now showing excessive tailing after only about 30 injections. (using a guard column connected by agilents ultra inert glass press fittings)
I am using a Shimadzu QP2010 Ultra with a glass insert liner with silanized glass wool. Thermogreen septa (Supelco). The splitless injection is done at 320°C liner temperature for 1.5 minutes at a pressure pulse of 300kPa at 70°C Column temperature. Solvent is n-hexane. I am using a constant flow method of 50 cm/sec. Temperature ramps are optimized for benzo-<->benzo[k]fluoranthene- separation, with a maximum temperature of 340 °C. (quite fast if anyone is interested
Anyone an idea why my system is unstable? I would expect a higher life time of my columns and less changing of the liner/Guard column. Samples arent highly contaminated, with low matrix contamination.
Is it a problem that I am simultanously checking for PCB compounds?
I used auto-tune to check for an air leak, m/z 28 was in appropriate range...
I also fasten the nuts inside the GC periodically (using graphite/vespel ferrules).
Related to this: how do you clean your glass insert liners and which kind of glass wool do you use? (and how much (pictures appreciated))
<-- at start
<-- 24 injections later
<-- Directly after Liner change
<-- 17 injections later