Agilent 7000 triple quad max room temp
Posted: Fri Aug 26, 2016 7:14 am
by BMU_VMW
Does anyone know what the maximum room temperature is for operating a Agilent 7000 triple quad?
It is pretty hot in here (+35°C) and I guess that is why my ms is shutting down on his own.
Re: Agilent 7000 triple quad max room temp
Posted: Fri Aug 26, 2016 8:30 am
by tkubowicz
Hello
Please see document below:
http://www.agilent.com/cs/library/sitep ... ist_EN.pdf
Regards
Tomasz Kubowicz
Posted: Fri Aug 26, 2016 9:52 am
by Hornet
It's normal. When room temp is around 35°C the pump temp raises over 75-80°C and it shuts off to go into safe mode.
Posted: Fri Aug 26, 2016 7:07 pm
by James_Ball
We got that hot here in the lab last summer when the air conditioning broke and our 7000 kept going, but the room temp was too hot for the GC to cool back down to the 35c starting temp( the last room temp I read before leaving for the evening was 36c) . The humidity was so high that the floor looked like you just mopped it. I imagine we were right on the edge of it shutting down too.