Agilent 7000 triple quad max room temp

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Does anyone know what the maximum room temperature is for operating a Agilent 7000 triple quad?

It is pretty hot in here (+35°C) and I guess that is why my ms is shutting down on his own.
Hello

Please see document below:

http://www.agilent.com/cs/library/sitep ... ist_EN.pdf

Regards

Tomasz Kubowicz
It's normal. When room temp is around 35°C the pump temp raises over 75-80°C and it shuts off to go into safe mode.
Davide Balbo from Italy
We got that hot here in the lab last summer when the air conditioning broke and our 7000 kept going, but the room temp was too hot for the GC to cool back down to the 35c starting temp( the last room temp I read before leaving for the evening was 36c) . The humidity was so high that the floor looked like you just mopped it. I imagine we were right on the edge of it shutting down too.
The past is there to guide us into the future, not to dwell in.
4 posts Page 1 of 1

Who is online

In total there are 21 users online :: 1 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry