To my surprise, all of the buffers I tried seemed to fluoresce -- only water/MeOH seemed to be transparent.
In my silliness, I wasted a lot of time trying different buffers -- tris, phosphate, carbonate, etc. Glycine was the worst of all, completely saturating my detector at 10mM.
I finally gave up trying to achieve a separation, and am now not using a column at all -- just using the HPLC for its fluorescence detector since I don't have a fluorimeter.
The past is past, as they say, but where can I read more about the fluorescent properties of buffers, so that I don't make the same mistake again? Anything similar to UV cutoff tables available?