- Posts: 14
- Joined: Tue May 29, 2012 7:24 pm
My Shimadzu RID-10A is showing some periodic baseline drops recently. We just had a service call to change the lamp and the solenoid valve but the problem persists.
At first, I thought it might be the pump but the back pressure did not change through the occurrence. I notice the the lamp energy (new) drops, whenever the baseline drops
I purged the line everytime this happens, but it's practically no use. The problem occurs with or without the column, and it doesn't matter whether it's flowing through the sample or reference cell. Has anyone seen anything like this before? Could it be the photo-detector getting worn out?
http://www.tiikoni.com/tis/view/?id=502e324
http://www.tiikoni.com/tis/view/?id=991b5cd
Thank you so much for your time!! Thank you!
Best,
Jason