Amino Acids GC problem (2 Peaks for some stds)
Posted: Wed Feb 15, 2006 12:48 pm
by Pat Crowley
I am attempting to analyse the 20 amino acid standards on a 30m DB-5 column.
I am using 1ml BSTFA + 1%TMCS, 1ml pyridine and heating at approx 80-85 degrees for 1 hour. I am getting peaks for all the derivatives(though some almost co-elute). However my problem is that for four of the amino acid standards I am obtaining two peaks as follows:
Aspartic acid 6.2 mins and 9.75 mins
Glutamic Acid 11.4 mins and 13.36 mins
Asparagine 11.66 mins and 12.23 mins
Glutamine 13.36 mins and 13.87 mins
Would anyone have a suggestion as to what is happening with these four amino acids?
Thanks.
Posted: Fri Feb 17, 2006 6:47 pm
by Consumer Products Guy
GCMS (if you have it) can tell you whether you've fully derivatized all the -N-H sites; the C-O-H sites derivatize much easier.
Posted: Mon Feb 20, 2006 12:28 pm
by Pat Crowley
Thanks for the reply. Unfortunately I don't have GC-MS. However I will try higher temp/longer derivatization time to see if any or all amino acids reduce to one peak.
Thanks.
Posted: Tue Feb 21, 2006 10:04 am
by Tony
I'm afraid longer derivatization won't help with BSTFA. If you use MSTFA or MTBSTFA, you will find most of the aa's you've listed will elute as single peaks due to more complete derivatization; however, glutamine and asparagine will still give double peaks. If you increase heat or derivatization time, glutamine and asparagine will break down to glutamate and aspartate, respectively.
Posted: Tue Feb 21, 2006 8:49 pm
by Rick
Put initial oven temp higher