noisy baseline @210nm
Posted: Wed Feb 15, 2006 6:10 am
I have a problem of the Agilent 1100 HPLC. Currently I found the baseline is very noisy at 210nm. Once changed the wavelength to 220nm and above, the baseline was very flat and stable.
To find out the reason, the system was bypassed the column.
No matter what mobile phase is water or AcN, the baseline fluctates about 1mAU.
The Agilent engineer has changed another set of optical unit and flowcell, the problem exists.
But he discovered that the surrounding air flow (from aircon) affects the baseline. When he covered the detector with some paper, the baseline shows not so noisy.
Anyone encountered the similar problem? Why temperature or air flow affect the baseline seriously? Or is there any causes behind?