using HPLC for fish medicines
Posted: Tue Feb 14, 2006 9:16 pm
by Jealemy
I'm using HPLC to determine the concentration of certain compounds on fish medicines. The medicines contain a mixture of the following compounds (in low concentrations) and between 80-95% dextrose:
Nifurpurinol
Acrinol
Malachite green oxalate
Metronidazole
Diflubenzuron
Cupric sulphate
Acriflavine
Citric Acid
Now, I've already used a polyamine II column (Acetonirtile/water 75/25) but some compounds coelute. A C18 column does not work at all. I'm now trying to optimize the gradient to see if I can resolve that problem. Does anybody have some experience with these compounds or suggestions????
Thanks a lot
Posted: Wed Feb 15, 2006 8:38 am
by Einar Ponten
You seems to have promising results by using HILIC conditions. While you still may need some other selectivity the ZIC®-HILIC column is an alternative.
Since you are using a gradient, I recommend that you initially run this from 5% to 50% buffer and then later tune the interval.
Posted: Tue Feb 21, 2006 9:36 am
by Rafael Chust
Usually you can try to separate complex mixtures of polar compounds with phenyl reversed phase columns, since it increases the retention of polar compounds and have a specific selectivity for molecules with pi orbitals.