I guess you kinda exactly understand the problem.
It does mention that at the end of run that the results were obtained using an enahnced integrator.
So could you please email me the a "default Method.M" for my Chemstation .
My email ID is
rithrr@yahoo.com
I have the version 8.03 and we are planning to get the version 8.10
Waiting for your reply.
thanks
I actually think I might know what's going on here (amazing to think that I am somewhat knowledgeable about any software). I think you previously had been using the ChemStation "standard" integrator and that you/someone clicked "OK" on the monitor message "Update to the Enhanced Integrator?" which can appear. Unfortunately, if one does that, it's like the roach motel of software (once a roach goes in, it never comes out), you cannot return to standard integrator for THAT METHOD.M . You'd have to load up a previously-saved file of that METHOD directory, or load up one of your other methods that have been saved using the standard integrator, or maybe the software default method (I'm not in the lab now, don't remember if it's lc_xxx.M or def_lcx.M, something like that). After loading such, you can make your other modifications than save back to the original Method.M file. What version of Chemstation are you using? I believe once one gets to A.10 that ONLY the enhanced integrator is used, and we use enhanced on our version A.09 HPLC-MS. Otherwise, I really and honestly prefer the "standard" integrator tons better, and that's what we use. We caution our operators at the production facilities to NEVER update to the enhanced integrator should that question appear. If you can't get there, leave your E-mail address and I can E-mail a "default Method.M" in your Chemstation revision level.