UPLC method transfer: validation?
Posted: Sat Feb 11, 2006 10:36 am
we are currently using some USP methods for raw materials.
As stated we use a L1 (C18) column with the prescribed mobile phase, solutions, .....
WHen we fullfill the method requirements (eg resolution, plate count, %RSD on 6 injections, ...), we can use the method to analyze our raw materials without validating the method entirely.
If we want to change to Waters UPLC, we can still use a C18 column, but it will be much shorter, and a with a smaller internal diameter.
We can use the Waters calculater to transfer the method to UPLC, and the results are amazing, increased resolution with decreased run time.
But what about the validation of the method?
As UPLC is based on pure chromatographic principles, one can easy understand that if you use 1.7µm particles, you have more plates/meter as with 5-10µm as stated in the monograph. For the use of shorter columns, you have to adjust your gradient, based on known calculations. For the use of smaller internal diameters, you have to adjust your gradient again and also your flow rate to get the same lineair velocity.
From my point of view, the only thing which is really changing is the column chemistry, the other adjustments can be calculated.
The change of column chemistry isn't that bad, because the monograph stated an L1 packed column, which can be hundreds of packed columns.
So, if we full fill the method requirements for resolution, plate count, %RSD, ... Do we have to re-validate the whole method for UPLC?
Or just a comparison between the methods by injecting the same samples on HPLC and UPLC and showing equivalency?
Thanks for your help