Suppression of residual DMSO peak by API

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I'm currently fine-tuning a headspace method for, among other things, residual DMSO. In the process of doing spike recovery with different APIs, I noticed something odd:

Depending on the exact sample, the slope of my calibration curve may be reduced by up to 20%, and this is consistent across the range of 500-5000 ppm. None of the other residual solvents in my standard exhibit this behavior (well, ok, DMF does, but to a lesser degree). The linearity is not a problem. This is very repeatable.

Has anyone else encountered this?

Sure, you may be saturating the gas in your headspace with the heavier molecular weight higher boiling point solvents (like DMSO)

In other words you may be out of the linear range giving the composition of your matrix. Or it may take a longer time to reach equilibrium that is linear. The presence of the other solvents in your calibration mix can affect the response (the speed of a linear partition) of the DMSO. The API can also affect this equilibration.

Everything in a headspace vial can affect the partition equilibrium of any of the other components. Even the amount of water in an API can affect your linearity.

Document your results very carefully. You may be glad you did at a later time.

best wishes,

Rod
2 posts Page 1 of 1

Who is online

In total there are 49 users online :: 0 registered, 0 hidden and 49 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 49 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry