RT Variation
Posted: Tue Aug 09, 2016 3:00 pm
The following is the method I am using for analyzing Impurity F by HPLC.
Impurity F (by HPLC):
(Reference: General method 2.2.29 Liquid chromatography and product monograph 0931 as per Ph.Eur.)
Derivatisation solution:Prepare the solution immediately before use. Dissolve 250 µL of fluorodinitrobenzene in 25.0 mL of acetonitrile (HPLC grade).
Blank solution:To 5.0 mL of acetonitrile (HPLC grade)add 100 µL of triethylamine and 1.0 mL of the derivatisation solution. Shake well and heat at 60 °C for 30 min. After cooling dilute to 10.0 mL with acetonitrile (HPLC grade).
Test solution:
Prepare the solution immediately before use. Suspend 10.0 mg of the substance to be examined in 5.0 mL of acetonitrile (HPLC grade)and sonicate for 5 min. Add 100 µL of triethylamine and 1.0 mL of the derivatisation solution. Shake well and heat at 60 °C for 30 minutes. After cooling dilute to 10.0 mL with acetonitrile (HPLC grade). Filter, or centrifuge at 800 g for 5 min before use.
Reference solution:
Dissolve 1.0 mL of Metformin impurity F WS (corresponding to 36 mg of Dimethylamine Hydrochloride or 20.0 mg of dimethylamine base) in 100.0 mL of acetonitrile HPLC grade. Dilute 2.5 mL of the solution to 100.0 mL with acetonitrile HPLC grade. To 1.0 mL of this solution add successively 5.0 mL of acetonitrile HPLC grade, 100 µL of triethylamine and 1.0 mL of the derivatisation solution. Shake well and heat at 60 °C for 30 min. After cooling dilute to 10.0 mL with acetonitrile HPLC grade.
Column:
- Size: l = 0.125 m, Ø = 3 mm
- Stationary phase: spherical end-capped octadecylsilyl silica gel for chromatography R1 (5µm)Column: Merck LiChroCART Purospher® RP18e 125-3mm, 5µ column
- Temperature: 30 °C (if column oven is available)
Mobile phase:
- Mobile phase A: phosphoric acid R, water R (0.1:99.9 V/V);
- Mobile phase B: acetonitrile HPLC grade
Mobile phase gradient:
Time (min) Mobile phase A (per cent V/V) Mobile phase B (per cent V/V)
0 60 40
0-10 60 45
40 55
10-11 45 25
55 75
11-15 25 75
Flow rate: 0.7 mL/min.
Detection: Spectrophotometer at 380 nm.
Injection: 5 µL.
Identification of impurities: use the chromatograms obtained with the blank solution and the reference solution to identify the peak due to the impurity F derivative.
Retention time: Impurity F derivative = Between 5 and 6 minutes
System suitability: From reference solution:Resolution to beminimum 3.0 between the peak due to the impurity F derivative and the nearby eluting peaks due to the derivatisation reagent.
Specification: Impurity F:Not more than the area of the corresponding peak in the chromatogram obtained with the reference solution (0.05 %).
Now The problem is, i am getting RTshift in impurity F RT and the limit is 5-6 it usually goes above six.
Can you please let me know what I can do?
Test solution:
Reference solution:
Column:
Mobile phase:
0 60 40
0-10 60 45
40 55
10-11 45 25
55 75
11-15 25 75
Injection: 5 µL.