by
dcc » Mon Aug 08, 2016 3:40 pm
Sorry for the late reply, took the data of our column oven and the time + m/z segments:
Column oven:
Initial 60°C, hold time 15min, to run time 15mins
Ramp 1, increase 15°C/min, hold time 2mins, target 195°C, to run time 26mins
Ramp 2, increase 15°C/min, hold time 0, target 225°C, to run time 28mins
Ramp 3, increase 5°C/min, hold time 0, target 260°C, to run time 35mins,
Ramp 4, increase 20°C/min, hold time 10mins, target 300°C, to run time 47mins
SIM time segments and target ions
18-21mins, Naphthalene, m/z 128 and 127.1
21-24mins, Acenaphthylene, Acenaphthene, m/z 152.1, 153.1, 154.1
24-26mins, Fluorene, m/z 166, 165.1
26-27.1mins, Phenanthrene, Anthracene, m/z 176, 178.1
27.1-28.6mins, PCB28, PCB52, m/z 220, 255.9, 257.9, 291.9
28.6-31.25, Fluoranthene, Pyrene, PCB101, PCB118, m/z202.1, 203.1, 253.9, 325.9
31.25-33.75, BaA, Chrysene, PCB 138, PCB 153, m/z 226.1, 228.1, 289.9, 359.9
33.75-39, BbF, BkF, BaP, PCB180, m/z 250.1, 252.1, 323.9, 393.8
39-47, InP, DbA, BghiP, m/z 276.1, 277.1, 278.1
So yeah, it looks like it may be column bleeding, I guess. Chromatogram looked similar again, standards looked contaminated with 'ghost peaks'. A septa change "showed" an improvement. But could it be, that column bleeding shows up periodically or would it show in every chromatogram?