Agilent 1100 Baseline

Posted: Mon Aug 08, 2016 11:28 am
by jsciscoe
Hi, this is my first post here.
I am pretty new to Agilent HPLCs and we are experiencing some intermittent baseline issues. The ripple is very low (relatively flat baseline). However, it is shaped like many plateaus lined up side by side. The baseline will suddenly drop about 15 mAu, then remain for 10-20 seconds before returning to original position for ~30 seconds. It looks almost like a square wave. This continues throughout the runs. Pump pressure remains constant throughout this which leads me to believe this is possibly a detector issue. However, there is no degasser installed on this model, so I am wondering if that could be an issue as well. Has anything seen this before?

Re: Agilent 1100 Baseline

Posted: Mon Aug 08, 2016 2:29 pm
by KM-USA
However, there is no degasser installed on this model, so I am wondering if that could be an issue as well.
I would NEVER try to get "decent" results without a degasser. In a pinch, heat/stir pull a vacuum on your mobile phase and let it cool, but that's a band-aid.

So fix first what you know is "bad".

Re: Agilent 1100 Baseline

Posted: Mon Aug 08, 2016 4:22 pm
by JI2002
An alternative is to purge the mobile phase with equal volume of He. For example, purge 1 L of mobile phase with 1L He.

Re: Agilent 1100 Baseline

Posted: Tue Aug 09, 2016 11:42 am
by lmh
... but it's quite hard to see how a degassing issue would create a reliable and regular square-wave. What wavelength is this? Does it happen at all wavelengths, or just some? Does it correlate with pump-rate (i.e. if you pump half the speed, do the plateaus double in length?).

Re: Agilent 1100 Baseline

Posted: Wed Aug 10, 2016 6:56 am
by lpnian
Hi jsciscoe , what is your detector on your HPLC?

Re: Agilent 1100 Baseline

Posted: Wed Aug 10, 2016 5:21 pm
by jsciscoe
Thanks for the replies
The detector is a DAD and the irregular baseline appears both at 214 and 300 nm. Unfortunately, I am unable to consistently duplicate the issue so I have not been able to experiment with pump pressures, flow rates, etc for troubleshooting. This problem will arise for 5 or 6 samples once or twice a week and then disappear.

Re: Agilent 1100 Baseline

Posted: Thu Aug 11, 2016 2:26 am
by lpnian
Hi, if that, i suggest you can do diagnose of the performance of UV lamp, it seems the lamp is going to disable.

Re: Agilent 1100 Baseline

Posted: Wed Aug 17, 2016 5:14 pm
by jsciscoe
Quick update..
I replaced the lamp and so far, so good. Thanks for the help.