Kajjo » Wed Feb 01, 2006 1:52 pm
Well, that looks like the lower volatile compounds do not actually get onto the column. Most probably the injector does not heat properly, please check! If the injector temperature is far below boiling point of the analytes, they will not be evaporated and brought onto the column (more or less, at least). The decreasing intensity and peak broadening points clearly towards injector problems!
>> check liner is clean (free of dirt and septum particles) and injector heats properly
I figure from your reply that the high-temperature column bleeding is visible? If yes, both the MS and the oven temperature control are fine, so please have a short look on the chromatogram!
Also, the mass spectra of the high-temperature range should exhibit the typical polysiloxane peaks. That would ensure that the MS works. However, as you describe the problem, it is most likely not due to the MS.
By the way, but not relevant for your problem:
I would use smaller amounts than 2 ul for injection, if possible, since too large injection volumes increase peak width and decrease resolution. Of course this depends on your injector liner dimensions. About 0.5-1.0 ul should be fine, normally, given a split rato of approx. 1:25 to 1:75.
Kajjo