Greetings and thank you in advance. I am relatively new to GC-MS on my own, so any help is appreciated.
I have a Perkin Elmer GC-MS (Clarus 580 - SQ 8 S) with Turbomass software that was setup in January 2016 by a rep, but never tested. After correcting some cables, instrument and PC are interfacing and operating ... but results are poor (see picture links below). Eventually, will be studying tea extracts, so PI wants to use MeOH as a sample solvent (seem reasonable?). Injecting 1uL into a 200C injector, no split on injection, carrier gas at 2mL/min, split flow 75mL/min, running on a PE Elite 17-MS column (rated for operation between 40C - 300/320C). Tunes and calibrates well, system is leak free. As funds are limited, I certainly don't want to waste money while trying to resolve the issue ... as well as not introducing additional problems.
Screen Shots:
1) https://drive.google.com/open?id=0B8Z18 ... E9oaGpyLVU
Blank MeOH Sample. (Also injected an air sample with similar result)
2) https://drive.google.com/open?id=0B8Z18 ... WRqV1A4NHM
Used a sample of Propyl 4-hydroxy-benzoate, provided by PI
Please note: Top portion of each picture is scaled to column saturation at 1e10, and bottom portion is scaled to mid detection levels 1e8.
Primary problems:
1 & 2) RELATIVELY HIGH BASELINE & BLEEDING. Was told limit of detection is around 1e7, and saturation around 1e10. While temperature is below 140C, baseline itself is adequate, but it climbs up to 1e8 and 1e9 by 210C, and bleeds at higher temperatures (the blank shown, final temp was 275C; the sample shown, final temp was 200C). To get adequate signal in this scenario, I need to keep temperatures quite modest, and have very strong signals bumping up near saturation ... anything at e8 and lower has interference.
3) RANDOM PEAKS. Get constant blips in the baseline. While they become more frequent and more intense at higher temperatures, they are present throughout. Always similar masses: (73, 147, 221, 281, 355) at the lower temperatures, and (150, 253, 315, 393, 451) at higher temperatures. Library search usually matches to variants of siloxanes.
My interpretation is that all of the problems are all related to a column that has gone bad and is degrading. Constant shedding gives random peaks, poor absorption onto the column, sample required for acceptable signal:noise nears column saturation and causes broader peaks and tailing, and column degrades above 240C to cause bleeding. Is this reasonable? Any other obvious cause I should investigate? And solution is just to replace the column? Is there a way to recondition it to stop the shedding and save costs? Acceptable to operate under the current conditions (ramps under 20C/min, max temp 215C)?
Trying to understand why the column would've gone bad ... It was sitting around for almost 7 months with just the occasional injection. Age and minimal maintenance kill it? Is there a solvent/column compatibility issue? If column is rated to 320C, why does it bleed at 215C and sheds throughout? Ramps were within limits (5 - 20C/min) and max temperature used was at least 25C under column max.
Any advice ...? Just spend the 800 necessary for the new column? :/ Thank you again
I have a Perkin Elmer GC-MS (Clarus 580 - SQ 8 S) with Turbomass software that was setup in January 2016 by a rep, but never tested. After correcting some cables, instrument and PC are interfacing and operating ... but results are poor (see picture links below). Eventually, will be studying tea extracts, so PI wants to use MeOH as a sample solvent (seem reasonable?). Injecting 1uL into a 200C injector, no split on injection, carrier gas at 2mL/min, split flow 75mL/min, running on a PE Elite 17-MS column (rated for operation between 40C - 300/320C). Tunes and calibrates well, system is leak free. As funds are limited, I certainly don't want to waste money while trying to resolve the issue ... as well as not introducing additional problems.
Screen Shots:
1) https://drive.google.com/open?id=0B8Z18 ... E9oaGpyLVU
Blank MeOH Sample. (Also injected an air sample with similar result)
2) https://drive.google.com/open?id=0B8Z18 ... WRqV1A4NHM
Used a sample of Propyl 4-hydroxy-benzoate, provided by PI
Please note: Top portion of each picture is scaled to column saturation at 1e10, and bottom portion is scaled to mid detection levels 1e8.
Primary problems:
1 & 2) RELATIVELY HIGH BASELINE & BLEEDING. Was told limit of detection is around 1e7, and saturation around 1e10. While temperature is below 140C, baseline itself is adequate, but it climbs up to 1e8 and 1e9 by 210C, and bleeds at higher temperatures (the blank shown, final temp was 275C; the sample shown, final temp was 200C). To get adequate signal in this scenario, I need to keep temperatures quite modest, and have very strong signals bumping up near saturation ... anything at e8 and lower has interference.
3) RANDOM PEAKS. Get constant blips in the baseline. While they become more frequent and more intense at higher temperatures, they are present throughout. Always similar masses: (73, 147, 221, 281, 355) at the lower temperatures, and (150, 253, 315, 393, 451) at higher temperatures. Library search usually matches to variants of siloxanes.
My interpretation is that all of the problems are all related to a column that has gone bad and is degrading. Constant shedding gives random peaks, poor absorption onto the column, sample required for acceptable signal:noise nears column saturation and causes broader peaks and tailing, and column degrades above 240C to cause bleeding. Is this reasonable? Any other obvious cause I should investigate? And solution is just to replace the column? Is there a way to recondition it to stop the shedding and save costs? Acceptable to operate under the current conditions (ramps under 20C/min, max temp 215C)?
Trying to understand why the column would've gone bad ... It was sitting around for almost 7 months with just the occasional injection. Age and minimal maintenance kill it? Is there a solvent/column compatibility issue? If column is rated to 320C, why does it bleed at 215C and sheds throughout? Ramps were within limits (5 - 20C/min) and max temperature used was at least 25C under column max.
Any advice ...? Just spend the 800 necessary for the new column? :/ Thank you again