We're trying to develop a method to quantitate GRO in the range of 10ppm-500ppm in soils by methanol extraction and then headspace injection with FID via a Agilent Select Mineral Oil column.
I've tried the direct headspace method before (Extracting GRO directly from the soil in the headspace vial) and while the method worked flawlessly on standards and even with matrix spikes into clean soil, the method failed to meet recovery percentages in actual field samples, which is very odd indeed.
(Samples were double checked via GCMS method 8260 by calculating total voc values).
Qualititive perliminary tests using methanol extraction did work.
however, when trying to build a calibration curve my low concentrations don't seem to work because of methanol tailing into the retention time window.
(the method works by checking a retention time window, then summing the area in the window)
What can I do?
Currently my starting temperature is 40C for 4 minutes, then ramping up to 120C. My RTW is between 1.2 and 5 minutes, and my methanol tails to about 1.5. all within the 40C isotherm.
My split ratio is set to very low 1:1 in order to achieve the neccessary sensitivity.
I also need to mention that 10ppm 25ppm "look" like they would've worked if I had no methanol tail in the RTW, 50ppm already works fine, but the environmental regulator wants 10ppm to be the LOQ, they might be okay with 25ppm but I don't think they'll settle for 50ppm.