Strange Peak From Fluorescence Detector
Posted: Tue Feb 07, 2006 10:42 pm
Hello everyone.
I have a problem with the fluorescence detector from Waters that my company purchased a few months ago. It is connected in series (as a system) to a UV detector and 2690 Separations module. The flow from the 2690 is first through the UV detector (2487 dual wavelength absorbance detector) and, then, through the fluorescence detector (2475 Multi λ), which then leads out to waste.
I am running 100% MeOH as the mobile phase and using 100% MeOH as the test sample. I'm trying to minimize the variables within the run. The peak from the UV detector is as it should be, but the peak from the fluorescence detector is inverted (looks like an aqueous inversion). The two peaks (one from the UV detector and one from the fluorescence detector) are mirror images of each other.
What is going on? I've tried for a week to get rid of the inverted peak, but nothing seems to help. There is NO water running through the system.
The fluorescence detector's parameters are all defaults, except for the excitation and emission λ, which are:
ex λ: 254
em λ: 402
This is a recent development. It was fine until about a week ago. I've tried 3 different columns (all the same - from the same manufacturer). All 3 columns are:
Symmetry C18 RP
4.6 i.d.
75 mm length
3.5 um particles
Any help would be greatly appreciated.
ex λ: 254
em λ: 402
Symmetry C18 RP
4.6 i.d.
75 mm length
3.5 um particles