Crompack GC

Posted: Sat Feb 04, 2006 7:44 am
by chemengineer
Dear Friends,
First I would like to say hello to all. I am a chemical engineer and I am going to work with GC (Chrompack CP9001), I faced with a problem today, when I was starting to work with the instrument, it shows me "Check LN2" and the signal for B was going up and up !
Would you please help me to know what is LN2 ? I have checked its catalog and haven't found it.
Hope to hear from you.
Thank you,
Semiramis

Posted: Tue Feb 07, 2006 10:18 am
by mahs
hi

If you are using nitrogen is carrier gas, if so check the flow of carrier gas.

Re: LN2

Posted: Thu Mar 09, 2006 1:28 pm
by jwmol
LN2 stands for Liquid Nitrogen. On the 9001 this may be used for oven cooling. You may have the CRYO MODE selected on page 7 (options page) in combination with an oven temperature setpoint that is below ambient.

JW