RP-Amide HPLC column stability
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
3 posts
Page 1 of 1
Does anyone have experience with using a reversed phase Amide column? It seems that it takes a long time for the column to come ready for use - excessively long time required for conditioning, is this typical for this phase or should I be looking for something else. The mobile phase is 65:35 ACN:Water with a gradient to 100% ACN.
-
- Posts: 686
- Joined: Mon Apr 29, 2013 4:37 pm
Nope, I've not observed anything such you have experienced with either the RP-Amide or its superficially-porous cousin, the "Express" version, nor with the Bonus-RP from Agilent. Is your injection solvent "weaker" in elution strength than 65:35 ACN/water?
MattM
-
- Posts: 715
- Joined: Wed Aug 04, 2010 2:45 pm
I agree with Matt.
Gerhard Kratz, Kratz_Gerhard@web.de
3 posts
Page 1 of 1
Who is online
In total there are 19 users online :: 2 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], Semrush [Bot] and 17 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], Semrush [Bot] and 17 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science