- Posts: 38
- Joined: Fri Oct 28, 2005 1:44 pm
My problem is in attempting an absolute calibration with pure hydrogen, which also has implications for the relative calibration of H2/air I want to do. There is always a large amount of air (shows up as a non-resolved double peak at 5.5 min, where hydrogen is at 3.5 min) when injecting what should be pure H2. I'm sampling the H2 in a fill setup from a cylinder of UPC hydrogen at a pressure of 30 psi. The setup is leak-free (I tested it, and it holds 30 psi for at least hours), and it is filled from the tank and then purged out with a vacuum pump 3x consecutively before the final fill. I'm then sampling directly from a non-leaking septum adapter fitting attached to a quick disconnect fitting, using a Valco sampling syringe (the type with the push-button valve at the bottom). I close the syringe valve before removing it from the septum, and don't open it again until injection.
I tried injecting different volumes of 'pure' H2, eg 10 uL and 5 uL - the peak for H2 halves but the air peak does not halve (although it does vary a little, like from peak area of 490000 to 390000). Using an absolute calibration of air, it appears that air accounts for at least 20% of the injected sample. What on earth is happening here, where could this be coming from? Is air coming through the inlet septum when I inject the needle? Surely the dead volume in the needle is way too little to account for this much air. Also, the peak area for hydrogen for consecutive injections of equal volumes seems to vary more than it should, for example 10 uL injections can give peak areas between 2.0E7 and 2.5E7.
Any help is greatly appreciated.