Problem with resolution
Posted: Sat Jul 23, 2016 8:26 pm
Is the resolution peak in chromatogram from analysisi crude extract using hplc must has value 1,5 or greater?
Chromatography Forum is a public discussion group where you can post questions, news, or messages of interest to chromatographers everywhere, but you must be registered to participate (registration is FREE). If you are a registered user, please log in.
http://www.chromforum.org/
So, how accurate/precise do you need to be? What is the relative size of the closest interfering peak?So there is no single answer to your question. If you are collecting fractions and purifying a crude extract, you may be able to get away with less than 1.5 (especially if you will be doing further processing). On the other hand, if you are looking at trace components, then you may need much higher resolution in order to get acceptable accuracy (e.g., Rs of 4, or 5, or . . . ).