Two detectors giving different area counts

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi, my Varian 3600 GC with Star software was working good till a unathorized tech. came to fix problem with autosampler. Now I am getting different responses from two FID detectors. Under same conditions and methods for both, Detector A gives area counts same as before while detector B area counts are about 10 times higher than that for Detector A. What would have been changed.? Any help and suggestions would be appreciated.

10X

I would check the amplifier setting for each detector or the connections from the instrument into your data system. Sounds like a 1V output was changed to a 10V output.
I would add that if they had the same respnose before the autosampler tech worked on the unit, something happened to the autosampler.

Check for leaks if this is a purge and trap.

Check to see that the autosamplers are delivering the same amounts.

Can you make a manual injection?

Did you check the FID flows?
3 posts Page 1 of 1

Who is online

In total there are 106 users online :: 1 registered, 0 hidden and 105 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Amazon [Bot] and 105 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry