Is the resolution peak in chromatogram from analysisi crude extract using hplc must has value 1,5 or greater?
Resolution of 1.5 is often referred to as "baseline" resolution, but that term is somewhat misleading. If you have two equal-size, Gaussian peaks with resolution = 1.5, there will be approximately a 1% overlap between the peaks. Therefore, Rs = 1.5 was originally called "99% baseline resolution". With time, the "99%" part dropped off and people started referring to simply "baseline resolution".

So there is no single answer to your question. If you are collecting fractions and purifying a crude extract, you may be able to get away with less than 1.5 (especially if you will be doing further processing). On the other hand, if you are looking at trace components, then you may need much higher resolution in order to get acceptable accuracy (e.g., Rs of 4, or 5, or . . . ).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
If i want to determine the concentration from marker of the crude extract, what is the better resolution for the marker peak?
I'll repeat what I said before:
So there is no single answer to your question. If you are collecting fractions and purifying a crude extract, you may be able to get away with less than 1.5 (especially if you will be doing further processing). On the other hand, if you are looking at trace components, then you may need much higher resolution in order to get acceptable accuracy (e.g., Rs of 4, or 5, or . . . ).
So, how accurate/precise do you need to be? What is the relative size of the closest interfering peak?

If you can tolerate errors of 10% or more and the closest interference is comparable in size or smaller than your marker, then you might be able to get away with Rs of 1 or so. If the marker is much smaller than the closest interfering peak (say, a 1:10 ratio for example) then Rs of 2 would be a good target. If the marker is very small compared to the interfering peak (say, 100:1), then a resolution of 4 or more may be necessary. All of the above are crude estimates and assume that your peaks are well-formed and symmetrical. If the peak shape is non-ideal, then higher resolution will be required.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Oh i see,thank you
6 posts Page 1 of 1

Who is online

In total there are 4 users online :: 3 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot], Majestic-12 [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry