Nile red
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
8 posts
Page 1 of 1
-
- Posts: 357
- Joined: Wed Aug 03, 2005 1:15 pm
Wheels,
Here is a reference,
Simple liquid-chromatographic method for Nile Red quantification in cell culture in spite of photobleaching.
Lamprecht A and Benoit JP, J Chromatogr B Analyt Technol Biomed Life Sci. 2003 Apr 25;787(2):415-9.
JMB
Here is a reference,
Simple liquid-chromatographic method for Nile Red quantification in cell culture in spite of photobleaching.
Lamprecht A and Benoit JP, J Chromatogr B Analyt Technol Biomed Life Sci. 2003 Apr 25;787(2):415-9.
JMB
-
- Posts: 4
- Joined: Wed Jan 25, 2006 4:04 pm
Thanks for that. I had already tracked down that reference but it's only available for hard cash and, much as I love my job, I'm not inclined to pay my hard earned dosh straight back in the form of journal subscriptions.
If no one can help then I'll just have a play with it and bung a very dilute sample down the c8 on a gentle ETOH gradient and see what happens!
If no one can help then I'll just have a play with it and bung a very dilute sample down the c8 on a gentle ETOH gradient and see what happens!
-
- Posts: 1074
- Joined: Mon Aug 30, 2004 5:42 pm
Wheels,
Here is the basic information from that reference. I think you can extrapolate from here...
Separation was carried out on a reversed-phase RP-18 column (LiChrospher© 100, Merck, Darmstadt, Germany). The mobile phase was kept as simple as possible which consisted of an isocratic eluent of methanol-water 93:7. The flow-rate was maintained at 1.0 ml/min. The effluents were monitored for their UV-Vis spectra of Nile Red and were recorded between 200 and 700 nm. For quantification purposes, Nile Red was detected by absorption at 559 nm, samples of 50 ml being injected into the column. All chromatographic analyses were performed at room temperature (25 °C).
Here is the basic information from that reference. I think you can extrapolate from here...
Separation was carried out on a reversed-phase RP-18 column (LiChrospher© 100, Merck, Darmstadt, Germany). The mobile phase was kept as simple as possible which consisted of an isocratic eluent of methanol-water 93:7. The flow-rate was maintained at 1.0 ml/min. The effluents were monitored for their UV-Vis spectra of Nile Red and were recorded between 200 and 700 nm. For quantification purposes, Nile Red was detected by absorption at 559 nm, samples of 50 ml being injected into the column. All chromatographic analyses were performed at room temperature (25 °C).
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
I hope you meant "50 μL" .samples of 50 ml being injected into the column.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 4
- Joined: Wed Jan 25, 2006 4:04 pm
Thanks for the information guys that'll do nicely! Just one small question though - what sort of Nile Red concentrations were being detected?
-
- Posts: 1074
- Joined: Mon Aug 30, 2004 5:42 pm
Tom, yes it was 50 uL.
Wheels, you may find below the LOD concentrations and the linearity range tested...
The limit of detection (LOD), defined as the lowest concentration of Nile Red which can be detected with a signal-to-noise ratio greater than 7:1, was established by serial dilutions with decreasing concentrations of Nile Red. The LOD was 9 ug/l for Nile Red.
Nile Red-free DMEM buffer solutions were spiked with the appropriate volume of the working solutions to contain Nile Red at concentrations of 3, 15, 50, 200, 800 and 3500 ug/l
Wheels, you may find below the LOD concentrations and the linearity range tested...
The limit of detection (LOD), defined as the lowest concentration of Nile Red which can be detected with a signal-to-noise ratio greater than 7:1, was established by serial dilutions with decreasing concentrations of Nile Red. The LOD was 9 ug/l for Nile Red.
Nile Red-free DMEM buffer solutions were spiked with the appropriate volume of the working solutions to contain Nile Red at concentrations of 3, 15, 50, 200, 800 and 3500 ug/l
-
- Posts: 4
- Joined: Wed Jan 25, 2006 4:04 pm
Fantastic, thanks. Have tested it out and it works a treat.
8 posts
Page 1 of 1
Who is online
In total there are 5 users online :: 2 registered, 0 hidden and 3 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Semrush [Bot] and 3 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Semrush [Bot] and 3 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science