Hi All,
I have just bought a grob test mix and the chromatography I get is good except the 2,3-butanediol. For this I get 2 peaks (the 1st is half the size of the 2nd). They both have the same mass spectra. All grob tests published have one peak for the 2,3-butanediol. Does anyone have the same problem? I am running a RTX5 both on MS and FID and the problem is the same. Injection split 100:1 1mlmin helium 0.1ul inj 250degC.
Best Regards
WK

WK

They are the two diastereoisomers. To get one peak you either need to obtain one of the diastereoisomers or the meso- form. I didn't bother - I just used the twin peaks as my column performance reference point.

WK

I just realised you said you bought the mix. I made up my own Grob mix and could only blame myself. Where did you buy it ?
3 posts Page 1 of 1

Who is online

In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry