ljc » Fri Jan 20, 2006 1:45 am
If these are older units, you'd have to test them side by side, since they probably don't perform as new and their original specs could be meaningless. The HP1047 is HUGE and heavy due to the T-stabilizing heat sink in it. The Waters unit is smaller (if it's the one I used, thought it was a 410) and wasn't as stable as the HP1047. But I recently aquired the new Agilent RI detector and it blows 'em both away in every aspect (except cost I guess).
If you don't need very high sensitivity, it might not matter much which used one you get, but the HP1047 does have many contact closures to remotely actuate purging, autozeroing, etc. However, zeroing it out with the adjustments is somewhat tricky. Be sure you get a manual with the HP if you buy it, or download it from Agilent's website, since it's not at all intuitive how to balance it out with all the buttons and knobs ! BTW, I have been told that the HP1047 was actually OEM'd by ERMA.
p.s. Have you considered ELSD instead of RI for your application ?