ELSD - inefficient nebulisation?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
3 posts
Page 1 of 1
We are having problems with ELSD signal response reproducibility - possibly caused by baseline noise. On reading a guide to ELSDs, it is my understanding that all solvent should be evaporated, and therefore the amount of waste solvent fairly minimal. We have noticed that the amount of waste solvent is approx. the same as the amount used to run the samples. Does this suggest inefficient nebulisation, and therefore a fault with the nebuliser? Or is it usual for most/all of the nebulised solvent to be recondensed? Any help would be much appreciated.
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
OK, I'm retired now, used 3 different ELSD units in my career. I can't claim to be an expert in ELSD, just an experienced user.On reading a guide to ELSDs, it is my understanding that all solvent should be evaporated, and therefore the amount of waste solvent fairly minimal. We have noticed that the amount of waste solvent is approx. the same as the amount used to run the samples. Does this suggest inefficient nebulisation, and therefore a fault with the nebuliser? Or is it usual for most/all of the nebulised solvent to be recondensed?
No, the amount of condensate for us was considerably less than the mobile phase volume, always. The more aqueous the mobile phase, typically the higher the ELSD nebulizer temperature. We used nitrogen at about 2 liters per minute.
Have you tried higher ELSD temperature like 50 C higher, to see if performance was better, worse, or the same?
-
- Posts: 155
- Joined: Tue Sep 27, 2011 8:54 am
User of a recent Agilent ELSD here. Modern ELSDs run on lower spray chamber temperatures and sample a small part of the nebulized eluent, so it MIGHT be OK if you see a lot of condense. It depends on the eluent composition and ELSD model.
3 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: troutnut1 and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: troutnut1 and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science