- Posts: 50
- Joined: Sun Nov 24, 2013 6:27 am
I am using helium with a new water trap and oxygen trap, I have conditioned the trap in the 5890, I have conditioned the column, and I've run the system now for almost four days and water is still at 50% of peak 69. Oxygen and Nitrogen are below 2%, so I don't think I have a leak. The water signal doesn't change with the calibration valve open or closed, and it doesn't change when column flow is reduced from 1ml to .5ml. It certainly seems like the water must be in the MS, but where could it be hiding? Could water vapor diffuse into the aluminum casting over time?
The signal went down pretty rapidly the first day, to a little over the peak for 69, and then dropped to its current value late on the second day. Since then, it's been pretty stable. I notice that when I close the valve for the PFTBA, the signal at 69 goes away almost instantly, but the water remains.