NPD startup problem

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hi, I am new in NPD chromatography.

We bought used GC HP 5890 with dual FID and NPD detectors on it, and managed to start the instrument. FID works just fine. Meantime NPD gives baseline around 6-15 pA and does not respond on injection of nitro compounds. We are confident that gases are pure and have proper flow rates: H2-4mL/min, Air 100 mL/min, He (make up) – 20 mL/min. I am not familiar with a faulty NPD “symptomsâ€

The next step is tp replace the bead to make sure you have a good one. The detector is very sensitive to the bead condition.

Regards,
Mark
Mark

The next step is tp replace the bead to make sure you have a good one. The detector is very sensitive to the bead condition.

Regards,
Mark
Thank you, I will definitely try the new bead. But generally speaking, is this possible to have a non-zero signal with “deadâ€
We started it!!

The problem was in cracked gasket at the detector base. It was leaking. Thank you all for the help and my special thanks to Abe Sanoja from MLS technologies Inc.
4 posts Page 1 of 1

Who is online

In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry