Regardless I've tried to do some experiments to see if an internal standard can be used to 1- notice when this error occurs and 2- correct for it. Normally the analyates are quantified with an external standard and there is no real reason I need to use internal standard for that sort of quantification.
The internal standard (IS) is added to the extraction solvent in a known concentration. If I compare a set of samples (n=10) where I perform a dilution and don't perform a dilution the %RSD of the IS peak area is 2.4% for diluted samples and for undiluted stock sample 0.9%. This suggests that the dilution step introduces some variation. In larger sample sets I've seen the %RSD go up to about 6% for the IS peak area.
I'm trying to think of best way to develop a correction factor for dilution error. Internal standard in solvent at 100ug/mL concentration. One thing I've tried is every set of samples generate a response curve for the internal standard in that batch of solvent over a range that covers undiluted and diluted samples 10-100ug/mL. Get a linear regression y = xb. If I made a 5 x solution I should have 20ug/mL IS. So I calculate what the area of IS would be if the IS was exactly 20ug/mL and get an expected peak area value. I then divide the expected peak area value with the actual peak area value and obtain a correction factor. I then use this correction factor to correct peak areas of analyates.
Does this make sense? How can I prove whether or not using the internal standard is more accurate then not using it? Is there a simpler way?
Thanks for any help anyone can provide.