FID major failure
Posted: Fri Jul 01, 2016 1:05 am
Hi all,
My first time posting and it is mainly out of desperation - I hope you kind folks can provide some guidance since I'm at a loss.
I've used this FID for going on 5 months now and it's been working fine. Two days ago, while performing a run by gas-sampling from our manifold, the signal shot up to what appears to be the maximum output for the detector (it runs off the screen in my analysis software) followed by a drop down to 0 (not previous baseline, but 0 units). After this, the signal increased back up to about 10^8 and has decayed to about 10^7, where it is oscillating between 0.75-1.25x10^7. I've never seen anything like this, but figured it must've been contamination. We installed a new piece of swage line in the manifold connected to this GC-MS and perhaps it wasn't cleaned prior so when I sampled via sampling valve, I introduced it to the column. No changes to makeup or fuel for the FID were made.
Since then, I've cleaned the collector plate/insulating disks/jet by sonicating in soapy water (no change), I've turned off helium makeup to the FID (no change), I've ruled out air/H2 since another GC using the same tanks and lines is working fine, I've disconnected the column entirely (no change). Right now, the output is about 1000-1200pA (I hear normal is about 0-20pA) and not really showing any signs of going down in the short term. I've baked the detector to 400 degrees C and likely stripped the section of column in there and only saw a minimal drop in the output. Has anyone ever seen anything like this? The interconnect spring doesn't appear damage, no parts appear visibly soiled.
I'm at a loss. Agilent says all signs are pointing to something with the electronics, which would suck. Any help would be super-duper-mega appreciated. Thanks!
My first time posting and it is mainly out of desperation - I hope you kind folks can provide some guidance since I'm at a loss.
I've used this FID for going on 5 months now and it's been working fine. Two days ago, while performing a run by gas-sampling from our manifold, the signal shot up to what appears to be the maximum output for the detector (it runs off the screen in my analysis software) followed by a drop down to 0 (not previous baseline, but 0 units). After this, the signal increased back up to about 10^8 and has decayed to about 10^7, where it is oscillating between 0.75-1.25x10^7. I've never seen anything like this, but figured it must've been contamination. We installed a new piece of swage line in the manifold connected to this GC-MS and perhaps it wasn't cleaned prior so when I sampled via sampling valve, I introduced it to the column. No changes to makeup or fuel for the FID were made.
Since then, I've cleaned the collector plate/insulating disks/jet by sonicating in soapy water (no change), I've turned off helium makeup to the FID (no change), I've ruled out air/H2 since another GC using the same tanks and lines is working fine, I've disconnected the column entirely (no change). Right now, the output is about 1000-1200pA (I hear normal is about 0-20pA) and not really showing any signs of going down in the short term. I've baked the detector to 400 degrees C and likely stripped the section of column in there and only saw a minimal drop in the output. Has anyone ever seen anything like this? The interconnect spring doesn't appear damage, no parts appear visibly soiled.
I'm at a loss. Agilent says all signs are pointing to something with the electronics, which would suck. Any help would be super-duper-mega appreciated. Thanks!