We have a Stratum Purge and Trap with VOCARB trap. We are using with it a 7890GC-7000C MSD with volatile interface (20:1 split). There is also a water trap (condenstator) inside the instrument.
I make calibration standards in methanol. Then i add it to 40ml vials. If i add too much standard into the vial (like 500 ul or more), my retention times are not stable. For example fluorobenzene go to 6,8-6,9 -depends on methanol amount- instead of 6,6 min. The higher the retention time of my compound i see less different in retention times (like 1,2-dichlorobenzene). Also i get much smaller peaks.
I am using SIM/Scan with 25 to 250 mass, but i dont see any 31 ion in my chromatogram so the methanol shouldnt enter into the column.
I am just corious why it is happening if anybody could explain me, i would be grateful.