the standard I inject contains heptane and toluene in dichloromethane at concentrations of around 0.6ppm.
the problem is my system suitability fails for %rsd on the three standard injections. two areas will stay consistent and then one will decide to jump up or down usually giving me an rsd of above 10%.
this is a validated method so please don't ask me to increase injection volume or anything like that as I would not be allowed.
running on an agilent 6890N in splitless mode,
split time 1.5min, 25ml/min
7ml/min He constant flow,
injector temp is 245
detector temp is 300
program: 35 for 8min, ramp 15C/min to 290C, hold 25min
column is DB-5 30M x 0.53umx1um film thickness
just to point out the heptane and toluene peak areas will both change to the same degree.
any ideas why this happens or could be improved would be appreciated.