blank run (No injection)
Posted: Fri Jun 24, 2016 12:43 pm
by y.ALKHADIRI
As i know it's very important to always run an instrument blank as the first thing to do in the morning before anything else, if possible..
i need some help, i am using GC 7890B From Agilent Technology, my question is how to run a blanc analysis without any injection ?
Re: blank run (No injection)
Posted: Fri Jun 24, 2016 1:03 pm
by tkubowicz
Hello
1.Load your method
2.Go to "Run Control" tab
3.Open "Sample info"
4.Fill all columns and leave "location" empty
5.Click "Run"
System will run loaded method with no injection.
Regards
Tomasz Kubowicz
Posted: Fri Jun 24, 2016 1:12 pm
by y.ALKHADIRI
Thank you for your reply.
what if i want to add it in the end? . i mean is there any possibility to programme it automatically after a couple of injections ?
Posted: Fri Jun 24, 2016 1:13 pm
by y.ALKHADIRI
Hello
1.Load your method
5.Click "Run"
Regards
Tomasz Kubowicz
Posted: Fri Jun 24, 2016 1:30 pm
by tkubowicz
Hello
It depends on your software version. As I remember if you leave "location" empty in your sequence table you can run blank. But for some Chemstation version it doesn't work.
So create your sequence table and try to run it with "location" empty.
Regards
Tomasz Kubowicz
Posted: Fri Jun 24, 2016 2:25 pm
by dblux_
Hello
Regards
Tomasz Kubowicz
It works for ChemStation codename Leonardo, but does't worka for MSD ChemStation (codename Mustang).
Posted: Fri Jun 24, 2016 2:43 pm
by y.ALKHADIRI
Thank you so much
it works.
Hello
Regards
Tomasz Kubowicz
Posted: Fri Jun 24, 2016 3:04 pm
by James_Ball
Hello
Regards
Tomasz Kubowicz
For Mustang you would have to create a method without the autosampler to run at the end of a sequence. Not sure if you can set it to inject instantly without waiting for an external device or not but I think it is possible.
Posted: Tue Jul 05, 2016 6:32 pm
by duck duck goose
You could also set your injection volume in your sequence table to 0. This will cause the system to run a blank injection as the sequence table settings will override the method settings.
One other option would be to create a blank sample. Just fill a vial with your solvent and inject from that.