I would welcome anyones help on this if possible.

This is the scenario...........
HPLC
Mobile phase: Ethanol/water
Diluent: Ethanol

Known source of contamination in the diluent gives two seperate positive peaks.

If this contamination gets into the mobile phase and not the diluent, the same peaks are negative rather than positive.

I Understand that the negative peaks will be due to the absence of the contaminant at that RT but do not understand the mechanism behind it, i.e. what is happening on the column.

Thanks

When the contaminant is in the mobile phase, it establishes an equilibrium concentration in the stationary phase. When you inject a sample with a solvent different than the mobile phase, it alters the equilibrium temporarily. The negative peaks are "holes" in the equilibrium concentration in the stationary phase that propagate much like peaks.
Mark Tracy
Senior Chemist
Dionex Corp.
2 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry