Cleanest Headspace Vials
Posted: Mon Jun 20, 2016 2:25 pm
by Charles A. Burger
So I'm working with a Total Evaporation Technique Headspace method. It has become clear that Manufacturers are not very familiar with this technique because I am getting plenty of garbage in my chromatography.
One partial solution has been to bake out the vials at 240°C. This had helped, but I still do get extra peaks.
So has anyone else had dealings with this style of Headspace? If so, what vials and septum have you had the best luck with?
Re: Cleanest Headspace Vials
Posted: Mon Jun 27, 2016 11:19 pm
by gapeev
How about
1. baking glass at ~500C
2. Incubating plastic parts in conc H2SO4+NaCr2O7 briefly
Posted: Thu Jul 07, 2016 2:28 pm
by MSCHemist
Hmm I use total evaporative static headspace and don't have many issues. I see one or two triangular peaks occasionally (GC/MS no definative ID probably siloxane or similar) but otherwise pretty clean blanks (other than carryover up to 5% for polar compounds like small acids and alcohols). I use a Tekmar 7000 loop system and I reuse headspace vials. Nothing fancy just put them in my dishwasher. The 9ml vials are hard to get for that instrument. I use typical crimp caps.
Posted: Thu Jul 07, 2016 6:49 pm
What is your volume into the headspace vial. For me its 5µl of STD or Sample into a 10 ml Headspace vial and cap.
Posted: Fri Jul 08, 2016 1:59 am
by MSCHemist
7.5ul aqueous solution into a 9.2ml headspace vial. Tekmar 7000 takes an odd size.