HP 1050 Autosampler
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
11 posts
Page 1 of 1
We have a HP 1050 HPLC system. The system has been working great. This morning, we were booting up and the autosampler will not start. It turns on, then keeps the word initializing on the screen. None of the keys on the keypad will do anything, and we cannot get chemstation to load because it keeps trying to reach out and find the autosampler. Any advice would be greatly appreciated!
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Amy - I'd disconnect the communications cables from the 1050 autosampler, then turn it off and then back on, see if it goes to its ready state on its own. You also need to have your air supply to the rear connected and on, to initialize the pneumatics (can you hear the pneumatics operating, and see the circular tray go abck and forth?).
-
- Posts: 8
- Joined: Tue Sep 17, 2013 8:12 pm
Unplugged all communication cables and it will not initialize on its own. It does not make the pneumatic sound at all. It powers on, and says initializing on the screen, but does not start doing all of the motion that it should on startup. Air is plugged in to autosampler. Board problem perhaps?
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
What is the supply of air? Cylinder, house air? And checked that it's air is delivering correct pressure, and didn't just assume.....
-
- Posts: 8
- Joined: Tue Sep 17, 2013 8:12 pm
We have a dedicated air compressor to supply the instrument. Air pressure is coming in.
-
- Posts: 1
- Joined: Thu Mar 28, 2013 9:31 pm
I had this issue and figured I would post what I did to fix it. after checking the compressed air line and setting the autosampler control board to test mode and it not doing anything i removed and reseated every single board in the back. after this it worked. My guess was it was a bad connection with the firmware board or a replaceable chip on a board that was causing it to hang before the initialization test. The screen and keypad wouldn't respond but when you booted it you could hold control to go into self test mode. so try reseating/inspecting all the component boards.
-
- Posts: 154
- Joined: Tue Jan 16, 2007 10:44 am
Turn off the instrument. Eject all plug-in cards at the back (2-3 cm is enough) and push them back. Now try to start the instrument again.
OOps, Tjwillis777 told the same ...
OOps, Tjwillis777 told the same ...
-
- Posts: 4
- Joined: Sat Jun 07, 2014 7:52 am
Amy, had you repaired it yet? pls share the experience.
I have other troulbe with my 1050 autosampler and 100 vials tray (18596 M). I don't remember that I had reset something on the autosampler so It can not control the 100 vials tray now.
Anyone know how to repair it, please guide me.
Thank you.
P/S: 1050 HPLC is very good instrument. I have 4 and they are still work well but only one have 100 vials tray.
I have other troulbe with my 1050 autosampler and 100 vials tray (18596 M). I don't remember that I had reset something on the autosampler so It can not control the 100 vials tray now.
Anyone know how to repair it, please guide me.
Thank you.
P/S: 1050 HPLC is very good instrument. I have 4 and they are still work well but only one have 100 vials tray.
-
- Posts: 8
- Joined: Tue Sep 17, 2013 8:12 pm
I could not repair it. We ended up buying a replacement.
-
- Posts: 16
- Joined: Sat Dec 05, 2015 1:06 pm
Hello, my auto-sampler has the Not Ready light on. It initializes normally, the circular tray go back and forth. But I did not connect an air supply on the rear. Is it necessary to connect the air? Is this the reason why the Not Ready light is on?Amy - I'd disconnect the communications cables from the 1050 autosampler, then turn it off and then back on, see if it goes to its ready state on its own. You also need to have your air supply to the rear connected and on, to initialize the pneumatics (can you hear the pneumatics operating, and see the circular tray go abck and forth?).
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Yes, the air must be connected, the cylinder air or compressed air supply is what drives the movement of the tray and movement of the injector. We had ours plumbed to house compressed air at about 75 psi.Hello, my auto-sampler has the Not Ready light on. It initializes normally, the circular tray go back and forth. But I did not connect an air supply on the rear. Is it necessary to connect the air? Is this the reason why the Not Ready light is on?Amy - I'd disconnect the communications cables from the 1050 autosampler, then turn it off and then back on, see if it goes to its ready state on its own. You also need to have your air supply to the rear connected and on, to initialize the pneumatics (can you hear the pneumatics operating, and see the circular tray go back and forth?).
11 posts
Page 1 of 1
Who is online
In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 21 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 21 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science